light bulb current limiter
What size bulb do you use to fire up 100W amps?
			
			
									
									Owner/Solder Jockey Bludotone Amp Works
						Re: light bulb current limiter
Funkalicousgroove wrote:What size bulb do you use to fire up 100W amps?
Usually 60 watt. Bascially you only want to watch for brightness in the case of excessive power draw so any size will work.
Re: light bulb current limiter
60 watt is good general purpose (for 120v mains).
If you want to get all geeky, figure out the current, e.g.
60w/120v = 1/2A max
With the lamp in series, neither device is getting full main voltage (unless there's a bad fault in the DUT, in which case the bulb gets most/all of the volts and glows at near-normal brightness).
--mark
			
			
									
									
						If you want to get all geeky, figure out the current, e.g.
60w/120v = 1/2A max
With the lamp in series, neither device is getting full main voltage (unless there's a bad fault in the DUT, in which case the bulb gets most/all of the volts and glows at near-normal brightness).
--mark
- Funkalicousgroove
- Posts: 2235
- Joined: Mon Jul 25, 2005 8:04 pm
- Location: Denver, CO
- Contact:
Re: light bulb current limiter
Good thing,  that's what I ended up using before anyone responded, I guess Great minds think alike!!!
			
			
									
									Owner/Solder Jockey Bludotone Amp Works
						Re: light bulb current limiter
Are the grounds for the male/female plug ends directly connected in the current limiter? 
http://bad-domain/DumbleLite/BuildersGu ... 0Guide.htm
Normsters shows them going to ground (which you should be able to reach thru the male plug ground), but don't know if it's necessary/advanageous to connect to chassis directly.
			
			
									
									
						http://bad-domain/DumbleLite/BuildersGu ... 0Guide.htm
Normsters shows them going to ground (which you should be able to reach thru the male plug ground), but don't know if it's necessary/advanageous to connect to chassis directly.
Re: light bulb current limiter
I think that's just a drawing convention, there's no need for the mains ground connection to go anywhere except to the "output" socket.
It wouldn't hurt (from a safety point of view) to ground the box as well, if the box is made of metal.
--mark
			
			
									
									
						It wouldn't hurt (from a safety point of view) to ground the box as well, if the box is made of metal.